Fluke 8846A this is an advanced and high precision multimeter, that I do own and admire for it significant pack of features.
For an unknown reason, Fluke Corporation leaved as undocumented, the function at DC mode, of
VA high precision Power Meter mode.
Fluke 8846A at
DCI mode (primary measurement display), and due the activation of it
2ND Meas (second measurement display), this now measure
simultaneously DCV and DCI (due three wires connection with the load).
The
undocumented VA mode this is hidden within
2ND Meas additional menu option.
1) Note: Feel free to read 8846A User manual -> Measuring DC Current -> DCI/DCV.
----------------------------------------------------------------------------------------------
Fluke 8846A accuracy specifications for DC: 10 V range 0.0024 + 0.0005
100 V range 0.0038 + 0.0006
(three wire mode for VA,
+ 20mV burden voltage)
10 V range 0.0044 + 0.0005
100 V range 0.0058 + 0.0006
400 mA range 0.05 + 0.005
3A range 0.1 + 0.02
10A range 0.15 + 0.008
I am not an mathematician, but combined or total VA error, this is never greater than a
Class 0.2 DC Watt meter (at any combination of voltage and current range in use, at FLUKE 8846A).
----------------------------------------------------------------------------------------------
Why would some one to care about this detail?
In the application of power measurement, and for power efficiency calculation, the 8846A it can become your eyes at the DC side (Load), for a single DC rail measurements.
For the AC side (source), you will always need a modern single phase (bench top) power meter, Class 0.5 or better.
2) Note: Such a combination of instrumentation, it would warranty power efficiency calculation with no greater error of 1%
3) Note: 8846A include analog and digital filter, its one it can assist in favor of noise removal, and or VA averaging = a more stable measurement that you may trust.
It is well known that integration speed at DMM, this slow down at dual mode measurements.
But with the use of programmable DC load, and at very specific load conditions (10% 50% 80% 100%), these are stable loads and the performance of 8846A this it can not be challenged.