This personal journey started, by me reading one IEC guideline about digital power meter calibration.
Any single phase power meter / watt matter, this it should be inspected regarding power factor calibration, at the values of 1.000 and at 0.500 PF
My second though was of where I can find such an apparatus, that it can have adjustable Power Factor displacement?
Power Factor 1.000 = any resistive load and even a common light bulb.
For adjustable Power Factor displacement down to 0.500 we need another and a special load.
Next in line information that I came across, was about calibration inspection of energy meter by the use of phantom power.
And then I did seen a known to me circuit schematic, and this was not other than of my own
Variable autotransformer – VARIAC.
The in-depth electrical theory about variable autotransformer, it mention that at very light loads, bellow 10% or lower, the magnetizing current of the autotransformer will lagging power factor as low as 0.5 PF.
I did activate right away my Variac that is a
Powerstat 226U, 2KV model, from
Superior Electric brand in the USA.
My Powerstat 226U this includes a
NORMAL and a
BOOST (step-up mode).
I did connect Powerstat 226U as this to be a load, at my
GM86 (plug in power meter), with four digits PF resolution.
Superior Electric's
standby power specification for
Powerstat 226U (without load connected) this is approximately
9 Watt (230AC).
Originally I started testing NORMAL mode, this operating as INPUT voltage limiter.
By turning VARIAC shaft from minimum to Maximum I recorder bellow values.
Watt Minimum: 7.78W -- PF 0.601 -- 58 mA
Watt Maximum: 10.25W -- PF 0.685 -- 65 mA
The miracle started to happen as soon I activated the
BOOST mode = a much higher magnetizing current.
Watt Minimum: 10.61W --
PF 0.462 -- 96 mA
Shaft adjustment at: 11.62W --
PF 0.501 -- 100 mA
Watt Maximum: 14.23W --
PF 0.570 -- 106 mA
And here came my moment so
and I to scream Eureka (word), an exclamation for a discovery or invention.
In conclusion:I could never believe that power analysis at my own variable transformer, this helped me to understand of how valuable this is, as testing apparatus at single phase power meter calibration!!