So let's get started:
As I opened the package and placed the Metrahit Cal in my Hands for the first time it's been some of these "wow" experiences.
The Built quality is excellent and actually feels even a lot more rugged than my U1272A (and we know, thanks to Kiriakos, that this is a tough companion). The Metrahit Cal would fit more in the Category "Tank", but I wont give any references like that here
.
The rubber-gurd can be easily removed and feels sticky and firmly in your hands. So if you just use it in your Lab, you could remove that one and the Metrahit Cal is much smaller and more like a Multimeter.
The rubber guard let it feel quite big and chunky and it will take some space in your toolbag.
Big plus-point: The calibrator is running on four AA batteries. You can get them or some decent eneloops everywhere and don't need to worry.
The runtime is between 8h (as mA source) and 45h (mV and TC), which should be useable in the field. GMC claims it to have 0,05% basic accuracy on most modes. Just as RTD source this goes up to 0,1%, which is still fine.
Talking accuracy.... The Metrahit CAL comes DAkkS (formerly known as DKD) calibrated which has some tighter regulations than an ISO calibration. Nice to have, or as professional, you may even need a proper calibration due to legal regulations. As Hobbyist.....nice to have and it gives you some kind of trust.
The Display is pretty nice with big displays, but as a calibrator it isn't filled with too much information anyway.
If wanted, you could buy a (very expensive) USB cable and Software and control it from your PC and set up several Calibrators, as they can have a Network address. Well, don't spend the rediculous amout of money and set it up by hand.
Let's talk functionality:
The Voltage source has four ranges: 0-±300mV, 0-3V, 0-10V, 0-15V with an maximum current draw of 15mA. At 18mA the Calibrator will shut the sockets off to protect itself. You can remove the wires then and shut it on again and nothing happened.
The Frequency source goes from 1Hz to 1kHz in 1 Hz increments (feeling like a robot with all the 1's now) and has a fixes duty cycle of 50%. The amplitude can be set in the same range as the Voltage source (0-15V). So there is no pulse width modulation what is a bit sad and could be useful. For that purpose I still need my Function generator. The max. current draw is 15mA again.
The current source can work in three ranges: 4-20mA, 4-24mA and 0-24mA, where you can source up to 20V out of it. Well, it is a constant current source, so you'll have a limited load here (of course).
The Metrahit Cal can also work as current sink, what means it will lower your circuits current for about the specified current. Well, this is more or less a negative constant current source. You have the same ranges here like in the normal current source mode.
The resistance simulator can go from 5 to 2000Ohms in 0,1Ohm increments. You can apply a maximum source of 5mA through the Calibrator.
So as we already had the mV and resistance sourced, it is quite obvious that you could also source RTD's and Thermocouples. It can simulate a PT100 (-200 to +850°C), PT1000 (-200 to +300°C), Ni100 (-60 to +180°C) and Ni1000(-60 to +180°C). This can be pretty handy if you want to test a temperature control loop or so.
Thermocouples can be simulated in a variety of types. Look up the Manual for the ranges
.
Two nice features are the Interval setup, where you can specify a Range, the number of steps and duration of a step and the Calibrator runs these up and down. So you'll just need to watch the reading of the tested meter, logger, A/D converter or whatever. As you know, or should know, which steps will be used by the Calibrator.
The second mode, ramp, is a bit different as it isn't using steps, but a continuos ramp where you specify the time it rises, it stays on the top level and the time to lower again. Personally I prefer the Step function as it is more controllable than the ramp.
I have two scope patterns for you to show these modes.
So all in all...it can be very handy, but probably you don't need one :-D. It may help you when you design a lot of circuits or control loops or when you have to maintain that stuff. I just used it today to have a look on the A/D input on one of my arduino's.... as expected it's not top notch, but can be used at least
.
Today I got an email from my Chief, so it seems I will use it more often in the next weeks, as I will refurbish some equipment (Loggers and Sensors) and set them up in Taiwan next March. So I'll have fun with the Metrahit Cal within the next weeks
. It seems my Agilent's will have a trip home next year
.
I hope you don't mind the shit*y picture with small DOF and lots of moiré. It's been pretty dark and I needed to creep-in a lot onto my Display.